In the Automatic Pacer Testing Mode, measurements are performed semiautomatically on seven parameters: pacer rate, pacer width, pacer amplitude, pacing refractory interval, sensing refractory interval, sensitivity, and A-V delay. The pacer width measurement is combined with the PPM measurement. For each parameter being measured, 10 different test values are preprogrammed. The user has the option of running all 10 of the tests using default values, running one or more tests with user selected values, or eliminating any of the tests. User modified values, may be saved for use in subsequent tests as long as the DELTA 3000 is not turned off. The user can leave the modified value
unsaved so that the original default values will be output during subsequent tests. The test sequence continues until all of the parameters are tested. The DELTA 3000 stores the measured values of each test performed in memory. At the completion of the automatic test sequence, the results may be sent to a printer or to a PC through the RS232 interface. The set and DELTA 3000_User_Manual_R2.doc 34
measured values are compared. If there is a deviation of more than 10%, a comment to that effect is printed.